Quote:
Originally Posted by Chris K View Post
OK the 8601 is here It powers up. I think I might need a few more BNC cables! I tried to configure the instruments for the sweep characteristics test and after 4 hook-ups, I ran out!
I count four interconnection hookup cables

Blanking
Sweep
Sweep Inhibit
Aux out of the Sweeper tthe counter in rear.

In addition you need the generator RF out to apply to the circuit under test and a cable, typically scope probe to it's vertical input.

Have you verified that the HP8601a produces a CW output which frequency can be measured on the HP8600a?

Refer to the Operator check on pages 21 and 22 of the BAMA HP8601a manual pdf.

Last edited by Penthode; 08-20-2024 at 08:05 PM.